Manual CA Hanna checker?


shreek

Rekommenderade inlägg

- Turn the meter on by pressing the button, all segments will be displayed. When the display shows "Add", "C.1" with "Press" blinking, the meter is ready.

- Add 1ml of Reagent A to the cuvette using the syringe with the white plunger. Use the dropper to fill thecuvette to the 10 ml marl with deionized water and replace the cap. Invert 3-5 times to mix.

- Place the cuvette into the meter, close the lid and press the button. When the display shows "Add", "C.2" with "Press" blinking the meter is zeroed.

- Remove the cuvette from the meter and unscrew the cap. Add 0,1 ml of sample to the cuvetteusing the syringe with the green plunger.

Note: Foe improved accuracy fill the syringe to the 1 ml mark and dispense 0,1 ml by lowering the plunger to the 0,9 ml mark.

- Add 1 packet of Reagent B. Replace the cap and shake vigorously for 15 seconds or untill the powder is completely dissolved. Allow airbubbles to dissipate for 15 seconds before taking a reading.

- Place the cuvette back into the meter and close the lid. Press the button. The meter displays concentration in ppm of calcium.

(The meter automatically turns off after 10 minutes)

Länka till kommentar
Dela på andra sidor

Gå med i konversationen

Du kan posta nu och registrera dig senare. Om du har ett konto, logga in nu för att posta med ditt konto.

Guest
Svara på detta ämne...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Ditt tidigare innehåll har återskapats.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

×
×
  • Skapa Ny...